Redundant Ground MC #181423 10/07/08 02:06 PM
Joined: Sep 2003
Posts: 60
G
gunther Offline OP
Member
I haven't found any information about how to terminate the redundant ground wire that comes in the green mc that is suitable for patient care areas. One opinion was to fold it back over the mc and make sure the mc fitting clip caught it, another was to wrap it around the outside grooves of the mc and catch it with the clip, and another was to use land it under the bolt that tightens the clip.
My thought would be to just make it up with the regular ground in the device box.
Is there a detail on this proceedure that I am missing?

2017 / 2014 NEC & Related Books and Study Guides
Re: Redundant Ground MC [Re: gunther] #181424 10/07/08 02:36 PM
Joined: Dec 2000
Posts: 4,283
electure Offline
Member
Just fold it back, and cut it off at the armor, if you're talking about Southwire's product.

http://www.southwire.com/processGetPage.do?bp=featuredProduct.jsp

Click on "view installation steps"




Re: Redundant Ground MC [Re: electure] #181426 10/07/08 04:30 PM
Joined: Sep 2003
Posts: 60
G
gunther Offline OP
Member
Thanks, that helps a lot. Although the mc shown isn't the type suitable for patient care areas, with the insulated green ground and also the bare redundant ground that accompanies it, it does explain that the ground is sufficient because it is in contact throughout and evidently the fitting is sufficient for continuous ground purposes.
The only question I might have is that this type of mc is green in color and I wonder if it is a non-conductive paint. I appreciate your help.

Re: Redundant Ground MC [Re: gunther] #181427 10/07/08 04:47 PM
Joined: Sep 2003
Posts: 60
G
gunther Offline OP
Member
Update: I just called the manufacturer and they say that it has been tested to perform excellently when the bare conductor is cut off where it comes out of the jacket.

Re: Redundant Ground MC [Re: gunther] #181436 10/07/08 09:30 PM
Joined: Jul 2004
Posts: 625
S
SolarPowered Offline
Member
Just FYI, I believe that the HCF cable is actually AC cable, not MC.

Re: Redundant Ground MC [Re: SolarPowered] #181437 10/07/08 10:11 PM
Joined: Dec 2000
Posts: 4,283
electure Offline
Member
It's MC, a type that came out recently.

From the site linked to above:

"UL-listed and NEC-compliant MCAP and HCF MCAP can replace conventional AC or MC cable in a broad range of applications."

Last edited by electure; 10/07/08 10:27 PM. Reason: add some info
Re: Redundant Ground MC [Re: electure] #181438 10/07/08 10:21 PM
Joined: Jul 2004
Posts: 625
S
SolarPowered Offline
Member
According to http://www.afcweb.com/faq_.html

Quote
What's the difference between HCF and MC cable?

The difference between HCF and MC cables (as manufactured by AFC Cable Systems) is the grounding capabilities of the cable assemblies.

Type HCF cable has a green grounding conductor plus the armor/bond wire combination, thereby providing redundant or isolated grounding capability. HCF is a Type AC cable and is manufactured in accordance with U.L. 4.

Perhaps someone else manufactures an HCF cable that's type MC.

Re: Redundant Ground MC [Re: SolarPowered] #181440 10/07/08 10:34 PM
Joined: Jul 2004
Posts: 625
S
SolarPowered Offline
Member
Too late to edit my previous post: Now that I'm thinking about it, I do recall an announcement a couple years ago that somebody was making an MC version of HFC.

Re: Redundant Ground MC [Re: SolarPowered] #181824 11/02/08 12:11 PM
Joined: Mar 2005
Posts: 236
watersparkfalls Offline
Member
I know this has been answered already but,
the idea of grounding with the conductors wouldn't work for two reasons. 1-its aluminum and conductors are copper and 2-that bonding jumper is very thin(16awg ?) and breaks very easily...but diffrent metals cant be tied directly tgether.

H20

Re: Redundant Ground MC [Re: watersparkfalls] #181827 11/02/08 12:51 PM
Joined: Jul 2004
Posts: 9,534
G
gfretwell Offline
Member
The bonding conductor in the new MC cable is not that little strip you had in type AC. This is larger than the circuit conductors and in tight contact with the armor


Greg Fretwell

Featured:

2020 National Electrical Code
2020 National Electrical
Code (NEC)

* * * * * * *

2020 Master Electrician Exam Preparation Combos
2020 NEC Electrician
Exam Prep Combos:
Master / Journeyman

 

Member Spotlight
sbi
sbi
indiana
Posts: 47
Joined: March 2008
Show All Member Profiles 
Top Posters(30 Days)
BigB 3
Popular Topics(Views)
270,995 Are you busy
205,001 Re: Forum
193,118 Need opinion
Powered by UBB.threads™ PHP Forum Software 7.7.3