#212067 - 12/08/13 02:18 PM D type grounding?
Potseal Offline
Member
Registered: 02/23/13
Posts: 204
Loc: Saskatchewan
We're running power for medical equipment. The equipment comes with a service manual. In the service manual it mentions "D type grounding". Here is all the information as it is stated in the manual:

"Ground Terminal

Grounding resistance should be less than 100 ohms (D type grounding) ground terminal. Use a single ground to earth. Do not share with other electrical apparatus."


Never heard of "D type grounding". Based on the description it sounds like an "isolated ground" but since it's referring to the "ground terminal" I'm not certain. Anybody familiar with the term "D type grounding"?


Edited by Potseal (12/08/13 02:18 PM)
_________________________
A malfunction at the junction
Top
Work Gear for Electricians and the Trades
#212068 - 12/08/13 03:49 PM Re: D type grounding? [Re: Potseal]
electure Offline


Member
Registered: 12/24/00
Posts: 4259
Loc: Fullerton, CA USA

Could it by chance mean "dedicated ground"?
Top
#212069 - 12/08/13 06:52 PM Re: D type grounding? [Re: Potseal]
renosteinke Offline
Cat Servant
Member
Registered: 01/22/05
Posts: 5316
Loc: Blue Collar Country
The instructions appear to call for a ground that violates our code. Better have a chat with the inspector.
Top
#212070 - 12/08/13 07:12 PM Re: D type grounding? [Re: Potseal]
mbhydro Offline
Member
Registered: 02/21/03
Posts: 347
Loc: Winnipeg MB Canada
Never heard of that term either but looking at Google shows referenced to class A thru D grounding.

Some installation manuals that are in the search seem to be from Europe / Asia and have the line:
D type Grounding (grounding resistance is lower than 100 ohms)


Edited by mbhydro (12/08/13 07:14 PM)
Top
#212071 - 12/08/13 07:16 PM Re: D type grounding? [Re: Potseal]
gfretwell Offline


Member
Registered: 07/20/04
Posts: 9039
Loc: Estero,Fl,usa
They mean isolated ground. You run a dedicated, insulated conductor way the way back to the grounding electrode. You don't share "that conductor" with any other apparatus.

That commonly gets misunderstood to mean you drive a separate rod. That is not right.
_________________________
Greg Fretwell
Top
#212072 - 12/08/13 09:36 PM Re: D type grounding? [Re: Potseal]
Tesla Offline
Member
Registered: 06/16/04
Posts: 1273
Loc: Sacramento, CA
All grounding conductors have to be bonded to the neutral back at the SERVICE.

Truly isolated grounds have been tried -- and they ended up causing the whole dang structure to burn down to the ground during lightning strikes. (Florida)

It's essential to establish an equapotential voltage plane. That stops instantaneous arcs from jumping during lightning events -- or other weird transients -- such as a nearby transformer fault. (traffic collision)

_________________________
Tesla
Top
#212074 - 12/09/13 05:58 AM Re: D type grounding? [Re: Potseal]
Potseal Offline
Member
Registered: 02/23/13
Posts: 204
Loc: Saskatchewan
At some point I will meet the tech from the company who is supplying the equipment and he will likely confirm that "D type grounding" is in fact an "isolated ground". Until then, thank-you for the replies.
_________________________
A malfunction at the junction
Top
#212075 - 12/09/13 08:04 AM Re: D type grounding? [Re: Potseal]
ghost307 Offline
Member
Registered: 05/09/05
Posts: 900
Loc: Chicago Illinois USA
I have run into quite a few cases where the terminology ended up being the enemy.
I would suggest that you draw up what you want to provide and ask the tech to look at it. That works out better than agreeing on what to call it...only to find out later on that your definition differs from his.

Case in point; I have run across at least 3 meanings of the term 'dedicated circuit'. All of them being equally right (or equally wrong) since NEC does not define the term.
_________________________
Ghost307
Top
#212094 - 12/10/13 09:34 PM Re: D type grounding? [Re: ghost307]
Potseal Offline
Member
Registered: 02/23/13
Posts: 204
Loc: Saskatchewan
Originally Posted By ghost307
I have run into quite a few cases where the terminology ended up being the enemy.
I would suggest that you draw up what you want to provide and ask the tech to look at it. That works out better than agreeing on what to call it...only to find out later on that your definition differs from his.

Case in point; I have run across at least 3 meanings of the term 'dedicated circuit'. All of them being equally right (or equally wrong) since NEC does not define the term.


Excellent advice. Thank-you.
_________________________
A malfunction at the junction
Top
#212098 - 12/11/13 07:36 PM Re: D type grounding? [Re: Potseal]
HotLine1 Offline


Member
Registered: 04/03/02
Posts: 6776
Loc: Brick, NJ USA
Ghost307:
I agree that your suggestion is excellent, and well said.

My opinion of the OPs question, I would lean toward a EGC straight back to the source panel.

_________________________
John
Top
Page 1 of 2 1 2 >

Member Spotlight
Member Since: 11/17/00
Posts: 2232
New in the Gallery:
SE cable question
Featured:

2017 NEC and Related
2017 NEC
Now Available!

Shout Box


Who's Online
0 registered (), 52 Guests and 8 Spiders online.
Key: Admin, Global Mod, Mod
 
New in the Gallery:
SE cable question
 
Top Posters (30 Days)
Admin 47
HotLine1 43
gfretwell 19
Ruben Rocha 12
Trumpy 9
 
Newest Members
Freecrowder, clee512, Jdscott2005, FAIZAN, Regitest2

ECN Electrical Forums - sponsored by Electrical Contractor Network - Electrical and Code Related Discussion for Electrical Contractors, Electricians, Inspectors, Instructors, Engineers and other related Professionals